I'm fixing up one of these for a friend. Originally I replaced the meters, creating new scales a couple of years ago. Those were big round ones sitting outside panel. Then small modern ones fitted. Then he found a couple of SIFA 50uA meters similar in size to the original. I then made third and fourth set of scales. The ones I did first were not quite right.
So obviously he wanted all the calibration re-checked.
Anyway, at "5" on 0 to 10 scale of grid volts the voltages per range should be about 0.78, 2.6, 7.8, 26. Adjusting RV2 the pot goes to one end of travel to get not quite as high as 2.6, 7.8, 26 on three upper ranges, but then the lowest range is over 0.9V. You have to put pot to nearly the other end of travel to get 0.78V, then the other ranges are far too low, like about 2.4V and 23.5V.
The resistors on the range switch seem OK. We replaced the 60uF 250V smoothing cap on the Grid supply. We replaced the 10K wirewound pot as it was bad on the first third of the range, intermittent connection, so needles dancing. With the new pot the voltages are the same, except it's now stable on the lower third of the scale.
No doubt the 0.52 ratio to actual volts from scale setting is to do with the cunning way the AVO works. The voltage at the junction between RV2 and the resistor dividers etc is supposed to be -52V. Why is the ratio 0.52 rather than 0.5 or 0.707? I must re-read the Radiomuseum article.
I traced the schematic for nearly an hour last night examining how the AC modulation from the oscillator is added to the slider of grid voltage pot (RV4) on panel and gets to the thumbwheel switches / valve sockets. The final switches & sockets are not on the schematic.
I THINK I might have spotted the faulty part, I'll verify next time I go over. Probably next week.
Anyone else any ideas or encountered this issue of mismatch between 3, 10, 30 and 100 grid voltage ranges?
So obviously he wanted all the calibration re-checked.
Anyway, at "5" on 0 to 10 scale of grid volts the voltages per range should be about 0.78, 2.6, 7.8, 26. Adjusting RV2 the pot goes to one end of travel to get not quite as high as 2.6, 7.8, 26 on three upper ranges, but then the lowest range is over 0.9V. You have to put pot to nearly the other end of travel to get 0.78V, then the other ranges are far too low, like about 2.4V and 23.5V.
The resistors on the range switch seem OK. We replaced the 60uF 250V smoothing cap on the Grid supply. We replaced the 10K wirewound pot as it was bad on the first third of the range, intermittent connection, so needles dancing. With the new pot the voltages are the same, except it's now stable on the lower third of the scale.
No doubt the 0.52 ratio to actual volts from scale setting is to do with the cunning way the AVO works. The voltage at the junction between RV2 and the resistor dividers etc is supposed to be -52V. Why is the ratio 0.52 rather than 0.5 or 0.707? I must re-read the Radiomuseum article.
I traced the schematic for nearly an hour last night examining how the AC modulation from the oscillator is added to the slider of grid voltage pot (RV4) on panel and gets to the thumbwheel switches / valve sockets. The final switches & sockets are not on the schematic.
I THINK I might have spotted the faulty part, I'll verify next time I go over. Probably next week.
Anyone else any ideas or encountered this issue of mismatch between 3, 10, 30 and 100 grid voltage ranges?