12-04-2017, 10:47 AM
Today I replaced the following components.
VR935, VR937, CR935, C926, R926.
Replaced P126 3pin plug.
Turned on scope no panel lights on front no trace.
Turned off scope removed P126 3pin plug.
Turned scope back on, panel lights came on front of scope.
Primary voltages where:
-15v spot on.
+15v was good.
+115v was +116.7v.
+250v was 255.4v.
-75v was -73v initially then dropped millivolt by millivolt down to -70v I stopped recording then.
C926 tested good out of circuit but replaced with new cap.
I turned off scope quickly when the front panel lights never came on as I didn't want to smoke R926 again.
What else could cause R926 to smoke.